The correspondence article 'Where is the leak? '1 caught my eyes, not only because it is an eye opener to all of us but also reminded me of a similar incident.
Catheter mounts are of great use to add ‘that extra length’ to the breathing system when needed. Each catheter mount has a 22 mm female / 15 mm male end to connect to the breathing system and 15 mm female end to connect to the endotracheal tube or laryngeal mask airway. The length is around 15 cm.
While this provides the extra length required, it also increases the dead space by about 20 – 40 mls. Thus, I tend not to use catheter mount as much as possible to avoid an increase in dead space.
I anaesthetised an adult patient for a maxillofacial case, which required nasal intubation. I used a north facing preformed (RAE) tube for intubation. After transferring the patient to the operating table, I noted that catheter mount is necessary to avoid any drag and to prevent any chance of accidental extubation. When I requested for a catheter mount, I was presented with the one shown in picture.
In this, connectors at both end of the catheter mount were 15 mm female, which normally is attached to the endotracheal tube / laryngeal mask airway.
This was not an acute situation and we had time to use an alternative catheter mount. However, incidences like this in an emergency situation may be catastrophic, especially if this is the only one stocked in the trolley.
Thus, I totally agree and support the suggestions made by K K Ramaswamy, A Bogdanov and M Alcock1 on checking the integrity and leakage of catheter mounts immediately before use.
Dr M Lohit
Northampton General Hospital, Northampton,
E-mail: lohitm69@googlemail.com
Reference:
1. K. K. Ramaswamy, A. Bogdanov and M. Alcock Anaesthesia 2010; 65: 311.