We read with interest the study by Hamaekers et al [1] on equipment currently used for oxygenation during complete airway obstruction. The authors commented that physical properties of such devices have major clinical consequences and that alternatives may be safer. We feel that their laboratory work does not support this assertion and that the venture is contrary to current guidelines [2].
The study measured pressure within a 2 mm trans-tracheal catheter during attachment to insufflation devices. Catheter flow during intermittent insufflation was then assessed. The first two devices used three-way taps to change from in-line to a side-on configuration that substantially reduced catheter-tip pressure. The authors branded the pressure with the in-line configuration (the currently prevalent technique) as potentially dangerous, but did not study this further.
However, the relevant pressure is the pressure within the trachea itself. This will be substantially below the value of 71.1 cm H2O found within the catheter. Whilst egress down the catheter may be impeded, the risk of barotrauma was not quantified. Additionally, given that total upper airway occlusion is an extremely rare finding, elevated tracheal pressure from catheter insufflation is still likely to generate some pressure-release via the upper airway.
The side-on configuration used by the authors requires an additional section to join the catheter and tap Luer lock ports: the rarely seen male/male connector. This necessity is at odds with the need for simple, readily available equipment in the emergency situation.
The paper then presents timings for insufflation and for egress of 1000 ml of oxygen. No explanation is given for the choice of this large tidal volume or its clinical validity. Inclusion of either device increased egress time compared to an unconnected catheter. Egress time was longer with the side-on tap configuration than with the Oxygen Flow Modulator. No statistical comparison is offered for the respective minute volumes. This is surprising given the apparent efficiency of the Oxygen Flow Modulator for egress.
The authors conclude that modifying the in-line configuration of the three-way tap will make emergency jet ventilation safer. Their statement is not borne out by this study. Switching to a side-on configured tap is complicated by the need for a male/male connector. Switching to a novel device would be supported by it achieving a significantly superior minute volume across a range of clinically appropriate tidal volumes; this has yet to be demonstrated for the Oxygen Flow Modulator.
This study confirmed that bidirectional flow through a 2 mm trans-tracheal catheter in the laboratory is possible. Despite acknowledging the inevitability of hypercapnia in the clinical setting, the authors considered this a technique for slowly re-establishing oxygenation. Faced with an obstructed upper airway, guidelines indicate the need to convert to a definitive airway as soon as possible [2], and we suggest that the focus of management should be on minimising this duration.
L Cohen
H Morris
D Vaughan
Northwick Park Hospital,
Middlesex, UK.
No external funding and no competing interests declared
References
1. Hamaekers AE, Borg PA, Enk D. A bench study of ventilation via two self-assembled jet devices and the Oxygen Flow Modulator in simulated upper airway obstruction. Anaesthesia 2009; 64: 1353–8.
2. Henderson JJ, Popat MT, Latto IP, Pearce AC. Difficult Airway Society guidelines for management of the unanticipated difficult intubation. Anaesthesia 2004; 59: 675–94.